flow alarm with Bruker D8 Advance (installed 2008)

Forum for troubleshooting hardware problems
chegmr
Posts: 2
Joined: 06 Apr 2017, 09:24

flow alarm with Bruker D8 Advance (installed 2008)

Post by chegmr » 06 Apr 2017, 10:19

Since our chiller for our sealed tube and generator failed and we had a rental one installed, we have been having regular shutdowns with water flow alarms .
The water flow in D8 tools has always been above 4.5 l/min and when we watch it , it does not vary by more than 0.05 l/min

Is there a way in d8 tools that we can get a log of the water flow rate over time ?

I was able to do this with our Cryostream

We have replaced the flow meter just yesterday by one of our in-house engineers and the same error occurs so I am suspecting a board problem

Ive emailed Bruker but wondered if the community had experience of something like this.

Thanks for any ideas

chegmr
Posts: 2
Joined: 06 Apr 2017, 09:24

Re: flow alarm with Bruker D8 Advance (installed 2008)

Post by chegmr » 25 Oct 2017, 09:34

Eventually we found out the chiller was faulty not the diffractometer.
As a last resort we connected to the mains (which is strongly discouraged by the uni) and the problem went away, apart from the times when there is a problem with the pressure of our mains supply .
We are waiting on the installation of a building chiller system from which we have our own enclosed loop